Regius 110 Error - Servicetechniker Forum

Regius 110 error

Tags: PACS / DR / CR / Digital, KONICA

(23)
(14)

romanz

Regius 110 error

November 01, 2012 05:11

I need a service manual for Konica Minolta Regius 110 CR.
My problem is: error 06180 "open door and check inside"

Please help me.

Regards
Roman

5 people liked this
Log inor Register

to rate and post a comment

(50)
(4)

Douglas Gray

re: Regius 110 error

November 02, 2012 07:25

send me your email. douggray9@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

Bartosz Ogi

re: re: Regius 110 error

September 18, 2013 05:31

Hello. I need a service manual for Regius 110 too. I need type of erase halogen lamp. Regards. Bartek.

2 people liked this
Log inor Register

to rate and post a comment

Georgy Smith

re: re: re: Regius 110 error

October 07, 2013 04:19

write your e-mail

1 person liked this
Log inor Register

to rate and post a comment

(54)
(23)

ADRIAN GUTIERREZ

re: re: re: re: Regius 110 error

July 03, 2014 01:46

Please send me a copy of service manual regius110 to me too.

cmedica.gualdo@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(2)

Mahdi Madani

re: re: re: re: re: Regius 110 error

February 19, 2019 10:11

Please send me a copy of service manual regius110 to me too. thanks
madanimahdi1358@gmail.com

Log inor Register

to rate and post a comment

(1)

Ondrej Prochazka

re: re: re: re: Regius 110 error

August 14, 2014 08:50

Hello dear, I would like need type of erase halogen lamp. thank you.

Log inor Register

to rate and post a comment

(56)
(9)

Budi Gunawan

re: re: re: re: re: Regius 110 error

February 20, 2019 09:41

you want buy ? i have it

Log inor Register

to rate and post a comment

(82)
(7)
(6)

Alfred M. Brown

re: re: re: re: Regius 110 error

January 13, 2015 04:38

Could you Please send me a copy of the service manual for Konica Regus 110 to me too. alsr@precisionmedical.org

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: re: re: Regius 110 error

January 14, 2015 08:52

Sent to your address.
I need documentation of part numbers Regius 110. Do you?
greet
Roman

Log inor Register

to rate and post a comment

(13)

norman alfaro

re: re: re: re: Regius 110 error

February 12, 2016 09:01

Good day. can I have a copy of the service manual for regius 110? thanks. my email is normanalfaro03@gmail.com

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: re: re: Regius 110 error

February 12, 2016 03:41

Sent to your address.
greet
Roman

1 person liked this
Log inor Register

to rate and post a comment

(10)
(3)

Okunade Joseph

re: re: re: re: re: re: Regius 110 error

May 25, 2016 07:14

I have a Regius 110 with CS-2 which would not display a correct image as exposed on the cassette. It always show that its reading and after it finishes, a blank rectangle like shape is displayed on the image screen whenever I try to scan an exposed cassette. The machine and all connections seem to be okay, but could not display any image acquired. Could it be software problem or can someone please tell me what I should do to make the image display correctly?

Log inor Register

to rate and post a comment

(5)

Acti med

re: re: re: re: re: re: Regius 110 error

December 26, 2018 04:40

hi
please can you send me copy of service manual reguis 110 ?
thank you
actimed@menara.ma

Log inor Register

to rate and post a comment

(1)

zatovo jean

re: re: re: re: re: re: re: Regius 110 error

March 01, 2019 07:47

Hello,please can you send me copy of service manual reguis 110 ?
thank you at jean.andrianjatovo@yahoo.fr

Log inor Register

to rate and post a comment

imad hissam

re: re: re: re: Regius 110 error

May 03, 2016 04:34

hi georgy.

i need the service manual of REGIUS 110 HQ

Best regards

Log inor Register

to rate and post a comment

(22)

djamel amira

re: re: re: re: re: Regius 110 error

May 07, 2017 04:21

hI !if you got the manual ,can i have a copy .
thanks .
djamel.amira@hotmail.fr

Log inor Register

to rate and post a comment

(2)

Richard Pfarr

re: re: re: Regius 110 error

October 10, 2014 11:32

need halogen lamp part number

Log inor Register

to rate and post a comment

(2)

Richard Pfarr

re: re: re: Regius 110 error

October 10, 2014 11:32

rickp@onisite.com

Log inor Register

to rate and post a comment

(82)
(10)

Vu Toan

re: re: re: Regius 110 error

July 08, 2015 07:58

I can help you. Pls contact vutoan_medical06@yahoo.com

Log inor Register

to rate and post a comment

(2)

Mahdi Madani

re: re: re: Regius 110 error

February 19, 2019 10:11

hello. If you have received that file, please send me a copy please.
thanks a lot.
madanimahdi1358@gmail.com

Log inor Register

to rate and post a comment

(1)
(1)

antonella volta

re: re: Regius 110 error

September 20, 2013 09:02

hello, could you please send me the illustration of how to fix the error 06182 in our Konica Minolta Regius 110? we can't download the manual from internet!!!!
thank's in advance
antonella

1 person liked this
Log inor Register

to rate and post a comment

Georgy Smith

re: re: re: Regius 110 error

October 07, 2013 04:19

write your e-mail

Log inor Register

to rate and post a comment

(1)

Mário Mansinho

re: re: re: re: Regius 110 error

August 28, 2014 10:57

Good morning,
Could need some help for error "06080 - open door and check inside". I have the original manual but doesn´t help. Assistance only next week...Thanks!

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: re: re: Regius 110 error

August 28, 2014 06:16

Hello Mario
I need documentation of part numbers Regius 110. Do you?
I greet
Roman

Log inor Register

to rate and post a comment

(2)
(1)

majid ramezani

re: re: re: re: re: Regius 110 error

July 04, 2016 10:39

please lock and unlock the lock of optical unit under the optical unit

1 person liked this
Log inor Register

to rate and post a comment

(2)

Shakir Siddiqui

re: re: re: re: Regius 110 error

September 04, 2014 10:19

i need service manual for regius 110, my email id is: shakir.siddiqui@maroof.com.pk

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: re: re: Regius 110 error

September 05, 2014 09:39

I sent.
Regards
Roman

1 person liked this
Log inor Register

to rate and post a comment

(5)

Edgar Barrios

re: re: Regius 110 error

December 04, 2013 06:13

could i have the manual regius manual too ?

Log inor Register

to rate and post a comment

(2)

Shakir Siddiqui

re: re: Regius 110 error

September 04, 2014 10:19

Please send me Service Manual My Email Is: shakir.siddiqui@maroof.com.pk

Log inor Register

to rate and post a comment

(117)
(26)

Steven McMurray

re: re: Regius 110 error

December 19, 2014 10:15

Can you help me also

Log inor Register

to rate and post a comment

(5)

Henrik kjoelby

re: re: Regius 110 error

May 02, 2015 03:51

Hi Douglas
I have to go to a site where they have a Regius 110 with a lift problem, it moves Down but not up, the customer says something is broken.
It would be very helpfull to have the service manual so please send it to henrik5240@gmail.com
Henrik

Log inor Register

to rate and post a comment

Nanjibhai Chauhan

re: re: Regius 110 error

March 02, 2016 09:12

Hi I also need konica regius 110 service manual.

Log inor Register

to rate and post a comment

(14)
(1)

German Perez

re: re: Regius 110 error

March 11, 2016 11:05

good morning.

I need to buy a halogem lamp to regius 110hq ...can you sell it to me? I'm From colombia

1 person liked this
Log inor Register

to rate and post a comment

(1)

angel verma

re: re: Regius 110 error

February 12, 2017 12:06

hello- can you send me the Regius 110 service manual to reverseseoresearch@gmail.com ? much appreciated!

Log inor Register

to rate and post a comment

(4)

LUIZ ANDRE PENZIN

re: re: Regius 110 error

July 17, 2017 03:26

hI !if you got the manual ,can i have a copy .
thanks .
andrepenzin@gmail.com

Log inor Register

to rate and post a comment

(2)

Wolf Ofner

re: re: Regius 110 error

September 02, 2018 10:30

can you send me the Regius 110 service manual please?

Log inor Register

to rate and post a comment

khalil al hosn

re: Regius 110 error

November 04, 2012 12:38

Dear Roman.
I had the same problem, i solved it, The problem is in the Guide of the timing built, Its a plastic guide and Iam sure that its broken for that the sensor reads that there is an object inside since the timing built doesn't return to its initial point.
Hope these information are useful to you.
Best regards

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: Regius 110 error

November 05, 2012 03:15

Colleagues
servis manual 3.2.4:
06150-06182 Subscan Motor Errors
Description of the error: Encoder initialization error
Respoznse: 1. Confirm that the fixing shaft of the simple fixing is not lowered

And that was the cause: mechanical lock dropped Optical / Erase Unit, and the truck
was not going to starting position.
I picked the lock, the error resolved.
Thank you once again for your help.
Thanks Douglas Gray and khalil al Hosn
I greet
Roman

Log inor Register

to rate and post a comment

(4)
(1)

Roman Zaborski

re: re: Regius 110 error

November 05, 2012 03:15

Colleagues
servis manual 3.2.4:
06150-06182 Subscan Motor Errors
Description of the error: Encoder initialization error
Respoznse: 1. Confirm that the fixing shaft of the simple fixing is not lowered

And that was the cause: mechanical lock dropped Optical / Erase Unit, and the truck
was not going to starting position.
I picked the lock, the error resolved.
Thank you once again for your help.
Thanks Douglas Gray and khalil al Hosn
I greet
Roman

Log inor Register

to rate and post a comment

(9)

Mikimiki

re: re: re: Regius 110 error

October 07, 2013 12:45

Hi, could you send me a copy of the regius 110 service manual too? Thanks

Log inor Register

to rate and post a comment

Georgy Smith

re: re: re: re: Regius 110 error

October 07, 2013 04:19

write your e-mail

Log inor Register

to rate and post a comment

(9)

Mikimiki

re: re: re: re: re: Regius 110 error

October 07, 2013 06:04

moura.jmiguel@gmail.com
Thanks ;)

Log inor Register

to rate and post a comment

(5)

Edgar Barrios

re: re: re: re: re: Regius 110 error

December 04, 2013 06:13

could i have a copy of the manual too ? edgar.barrios@swxrayonilne.com

Log inor Register

to rate and post a comment

Georgy Smith

re: re: re: re: re: re: Regius 110 error

December 18, 2014 03:19

not working e-mail...

Log inor Register

to rate and post a comment

(2)

Mike Welsh

re: re: re: re: re: re: re: Regius 110 error

June 15, 2015 04:21

Could you please email me a copy of the Regius 110 manual? Mwelsh@emh.org. Thanks.

Log inor Register

to rate and post a comment

(5)

Edgar Barrios

re: re: re: re: re: re: re: Regius 110 error

September 04, 2015 12:57

I never saw your email and now i'm seeing that i typed my email wrong. edgar.barrios@swxrayonline.com
please let me know if you can send me the service manual for regius 110 i need to change the ip address on it. Thank you ,

Log inor Register

to rate and post a comment

(5)

Edgar Barrios

re: re: re: re: re: re: re: Regius 110 error

September 04, 2015 12:57

correct email is edgar.barrios@swxrayonline.com
Thanks i need to change the ip address on that regius 110.

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: Regius 110 error

November 05, 2012 03:13


Colleagues
servis manual 3.2.4:
06150-06182 Subscan Motor Errors
Description of the error: Encoder initialization error
Respoznse: 1. Confirm that the fixing shaft of the simple fixing is not lowered

And that was the cause: mechanical lock dropped Optical / Erase Unit, and the truck was not going to starting position.
I picked the lock, the error resolved.
Thank you once again for your help.
Thanks Douglas Gray and khalil al Hosn
I greet
Roman

Log inor Register

to rate and post a comment

(4)
(1)

Roman Zaborski

re: Regius 110 error

November 05, 2012 03:15

Colleagues
servis manual 3.2.4:
06150-06182 Subscan Motor Errors
Description of the error: Encoder initialization error
Respoznse: 1. Confirm that the fixing shaft of the simple fixing is not lowered

And that was the cause: mechanical lock dropped Optical / Erase Unit, and the truck
was not going to starting position.
I picked the lock, the error resolved.
Thank you once again for your help.
Thanks Douglas Gray and khalil al Hosn
I greet
Roman

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: Regius 110 error

November 06, 2012 08:53

Sorry, too many times pressed the mouse.
I greet
Roman

Log inor Register

to rate and post a comment

(50)
(4)

Douglas Gray

re: Regius 110 error

November 06, 2012 01:52

Would you have a manual for Kodak DR9000?

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: Regius 110 error

November 08, 2012 10:26

I'm sorry, I checked and unfortunately I do not have any documentation to DR9000.
I greet
Roman

1 person liked this
Log inor Register

to rate and post a comment

(6)
(2)

ARMANDO AGARIJO

re: Regius 110 error

December 18, 2014 10:03

Hi, could you send me a copy of the regius 110 service manual too? my email is armandoagarijo@hotmail.com
Thanks

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: Regius 110 error

December 18, 2014 03:19

I sent. I hope that helps.
Best regards
Roman

2 people liked this
Log inor Register

to rate and post a comment

(3)

EwKa

re: re: re: Regius 110 error

February 11, 2015 09:33

Hi, could you send me a copy of the regius 110 Konica Minolta service manual too? My Email is: 12ewakap@gmail.com
Thanks

Log inor Register

to rate and post a comment

(3)

EwKa

re: re: re: Regius 110 error

February 11, 2015 09:31

Hi, could you send me a copy of the Konica Minolta Regius 110 service manual too? My Email is: 12ewakap@gmail.com
Thanks

Log inor Register

to rate and post a comment

(3)

EwKa

re: re: re: Regius 110 error

February 11, 2015 09:37

Hi, could you send me a copy of the regius 110 service manual too? my email is 12ewakap@gmail.com
Thanks

Log inor Register

to rate and post a comment

(5)

Henrik kjoelby

re: re: re: Regius 110 error

May 02, 2015 03:51

Hi Romanz
I have to go to a site where they have a Regius 110 with a lift problem, it moves Down but not up, the customer says something is broken.
It would be very helpfull to have the service manual so please send it to henrik5240@gmail.com
Henrik

Log inor Register

to rate and post a comment

(117)
(26)

Steven McMurray

re: Regius 110 error

December 19, 2014 10:15

I have the Manuals

Log inor Register

to rate and post a comment

(5)

Henrik kjoelby

re: re: Regius 110 error

May 02, 2015 03:51

Hi Steven
I have to go to a site where they have a Regius 110 with a lift problem, it moves down but not up, the customer says something is broken.
It would be very helpfull to have the service manual so please send it to henrik5240@gmail.com
Henrik

Log inor Register

to rate and post a comment

(2)

Charles Davis

re: re: Regius 110 error

May 21, 2015 01:12

Could I please get a copy of the manual too? I have a 110 with a 06602 error (fans) and I happen to have an old 110 with I might be able to steal parts from.
Thanks,
Charles
daviscl1970@gmail.com

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: Regius 110 error

May 21, 2015 09:02

I sent, check e-mail,
greet
Roman

1 person liked this
Log inor Register

to rate and post a comment

(2)

Charles Davis

re: re: re: re: Regius 110 error

May 21, 2015 04:12

Thank you so much! This helps a lot.

Log inor Register

to rate and post a comment

scott kadechuk

re: re: re: re: Regius 110 error

June 08, 2015 01:15

Hello Roman,

Could you send me the Regius 110 service manual?
skadechuk@gmail.com

Thanks

Scott

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: re: re: Regius 110 error

June 09, 2015 08:44

I sent, check e-mail
greet
Roman

1 person liked this
Log inor Register

to rate and post a comment

(2)

Mike Welsh

re: re: re: re: Regius 110 error

June 12, 2015 03:43

Could you please send me a copy of the Regius 110 service manual? Mwelsh@emh.org. Thanks

Log inor Register

to rate and post a comment

(2)

Steve Layton

re: re: Regius 110 error

February 11, 2016 07:27

Hi Steven,
Could I get a copy of the Regius 110 manuals?
And where could I get some parts ?
SLXRAY@BELLSOUTH.NET
Thanks,
Steve

Log inor Register

to rate and post a comment

(82)
(10)

Vu Toan

re: Regius 110 error

September 16, 2015 08:02

I have muanuals of regius110
Pls contact me vutoan_medical06@yahoo.com

Log inor Register

to rate and post a comment

(2)

Steve Layton

re: re: Regius 110 error

February 11, 2016 07:27

Hi Vu,
I have an error 6300 on a Regius 110. Could I get a copy of the manual? Does it include a parts breakdown?
SLXRAY@BELLSOUTH.NET
Thanks,
Steve

Log inor Register

to rate and post a comment

(5)
(2)

devlim

re: re: Regius 110 error

June 01, 2016 10:16

I also have a problem with a Kon-Min Regius 110 Scanner.
No error code but msg: Device Error CR JM Error. Justifier Sensor NG (fixed Action)

If you have the service manual, I certainly would appreciate it if you will e-mail me a copy,
Thanks
Case@canxray.com

1 person liked this
Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: Regius 110 error

June 01, 2016 01:08

I sent, check e-mail
greet
Roman

1 person liked this
Log inor Register

to rate and post a comment

mehmet ali

re: re: re: re: Regius 110 error

December 14, 2017 06:54

my email adress is malibayram91@gmail.com
can you send me service manuel

Log inor Register

to rate and post a comment

oscar sotelo

re: re: re: re: Regius 110 error

June 10, 2018 09:16

Hi! I have the same error as devlim, can you send me the service manual, please?

oscar.spwn@gmail.com

Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: re: re: Regius 110 error

June 11, 2018 10:26

I sent, check emil.
Regards
Roman

Log inor Register

to rate and post a comment

(1)

Eric Chen

re: re: Regius 110 error

July 11, 2016 09:56

Hi Vu,

I have an error message "SENDING IMAGE / CHECK LAN
CONNECTION" occurred sometimes. I checked network cable related setting, but the error still occurred occasionally. Would you send me a copy of Regius 110 service manual? my email is eric041227@gmail.com.

Thanks,
Eric

Log inor Register

to rate and post a comment

(1)
(1)

Lucio

re: re: re: Regius 110 error

December 27, 2016 10:17

Good Morning.

Could you send me the service manual for the CR Regius 110, am I having the defect error 06182?
Follow my email luciosousa2803@gmail.com

Grateful!

1 person liked this
Log inor Register

to rate and post a comment

(23)
(14)

romanz

re: re: re: re: Regius 110 error

December 28, 2016 09:01

I sent you to emila
Regards

Log inor Register

to rate and post a comment

(5)

Zeynettin Karaaslan

re: Regius 110 error

May 05, 2017 10:30

Hello.
Does Regius 110 have service manuel in hand?

Log inor Register

to rate and post a comment

(5)

Zeynettin Karaaslan

re: Regius 110 error

May 05, 2017 10:30

biyomedikalege@gmail.com

Log inor Register

to rate and post a comment

(5)

Zeynettin Karaaslan

re: Regius 110 error

May 05, 2017 10:31

Hello.
Does Regius 110 have service manuel in hand?

Log inor Register

to rate and post a comment

(4)

LUIZ ANDRE PENZIN

re: Regius 110 error

July 17, 2017 03:27

hI !if you got the manual ,can i have a copy .
thanks .
andrepenzin@gmail.com

Log inor Register

to rate and post a comment

(4)

LUIZ ANDRE PENZIN

re: Regius 110 error

July 17, 2017 03:27

hI !if you got the manual ,can i have a copy .
thanks .
andrepenzin@gmail.com

Log inor Register

to rate and post a comment

(1)

Alex Mejia

re: Regius 110 error

September 09, 2017 11:16

Please can you send me a copy of the manual, Thanks in advance, alexmejiahn72@gmail.com

Log inor Register

to rate and post a comment

ifwat zamir

re: Regius 110 error

January 14, 2018 09:57

ameamir2013@gmail.com please send me your service manual...

Log inor Register

to rate and post a comment

ifwat zamir

re: Regius 110 error

January 14, 2018 09:57

please send manual ameamir2013@gmai.com

Log inor Register

to rate and post a comment

(3)

Carlos Gonzales

re: Regius 110 error

April 10, 2018 08:51

Ca you please send me the service manual.

Mail is a.magnani.g@gmail.com

Thanks in advance

Best regards

Log inor Register

to rate and post a comment

(1)

Tsutomu Sakata

re: Regius 110 error

April 24, 2018 09:57

Hello, I also need a service manual of Regius 110 and Unitea workstation. Could I please have a copy too?
harada@e-mcast.co.jp

Log inor Register

to rate and post a comment

(2)

Aytac SEVEN

re: Regius 110 error

May 01, 2018 05:05

This error message can be an encoder initialization error.

Log inor Register

to rate and post a comment

(25)
(5)

JOSE GUADALUPE CAMPOS

re: Regius 110 error

October 01, 2018 09:24

I have the same problem. How did you solve it?

Log inor Register

to rate and post a comment

(1)

Thania Soto

re: Regius 110 error

March 26, 2019 03:55

I also have a problem with a Kon-Min Regius 110 Scanner.
No error code but msg: Device Error CR JM Error. Justifier Sensor NG (fixed Action)

If you have the service manual, I certainly would appreciate it if you will e-mail me a copy,
Thanks

apithania@gmail.com

Log inor Register

to rate and post a comment

(1)

Mahmut Baykara

re: Regius 110 error

May 15, 2019 10:16

Please send me a copy of service manual
mahmutbaykara6556@gmail.com

Log inor Register

to rate and post a comment

THE (LEADER) IN MEDICAL IMAGING TECHNOLOGY SINCE 1982. SALES-SERVICE-REPAIR

Special-Pricing Available on Medical Displays, Patient Monitors, Recorders, Printers, Media, Ultrasound Machines, and Cameras. This includes Top Brands such as SONY, BARCO, NDS, NEC, LG, EDAN, EIZO, ELO, FSN, PANASONIC, MITSUBISHI, OLYMPUS, & WIDE.

advertisement

 

RSTI Newsletter

 

 

Hitachi