Francesco Didonato

OEC9900

May 13, 2019 10:39

Hello i have problem with oec 9900 at boot touch-screen don't working when i touch it and of tra to fluoro after fluoro COMUNICATION error somebody can help me about this problem?
Thank's

Log inor Register

to rate and post a comment

(9)

ALLEN GODFREY

re: OEC9900

May 15, 2019 10:25

I would start by checking the usb cable and connections from the touch screen monitor to the GPOS CPU. GPOS is the cpu with the usb extender on it. the touch screen should be one of the cables plugged into the cpu onboard usb jacks. I might be more helpfull if I have more details after you try this. arg4team8@bellsouth.net