Lorad Miv - System Error 33 And Bucky Error 51 - Servicetechniker Forum

Lorad MIV - System Error 33 & Bucky Error 51

(130)
(32)

FranciscoG

Lorad MIV - System Error 33 & Bucky Error 51

April 15, 2014 03:47

Dear friends,

I've got a Lorad MIV having a System error 33! and sometimes this other error Bucky error 51! does not allow me to make an exposure. Does anyone have any idea of what could this be or have the service manual or troubleshooting manual for this system that I can get, would be very appreciated.
Thanks.

1 person liked this
Log inor Register

to rate and post a comment

(3)
(1)

Jeff Sirk

re: Lorad MIV - System Error 33 & Bucky Error 51

April 16, 2014 08:31

your grid drive motors need to be replaced inside the bucky

2 people liked this
Log inor Register

to rate and post a comment

(2)

Eric Hazen

re: re: Lorad MIV - System Error 33 & Bucky Error 51

March 25, 2015 03:08

If I only get the error 33 is it still the motors or would the IR pcb be more likely?

Log inor Register

to rate and post a comment

(39)

Anthony Alleyne

re: re: Lorad MIV - System Error 33 & Bucky Error 51

October 07, 2019 10:33

Hello
does any one have a schematic diagram in pdf format for the lorad miv?

Log inor Register

to rate and post a comment

(27)
(3)
(5)

James McGinty

re: Lorad MIV - System Error 33 & Bucky Error 51

April 17, 2014 06:02

Bucky problem... usually motor, sometimes circuit board...

1 person liked this
Log inor Register

to rate and post a comment

(39)

Anthony Alleyne

re: re: Lorad MIV - System Error 33 & Bucky Error 51

July 30, 2014 09:30

i also need assistance with bucky

Log inor Register

to rate and post a comment

(312)
(22)
(8)

Joseph Jenkins

re: re: re: Lorad MIV - System Error 33 & Bucky Error 51

October 16, 2020 10:31

hello ..do you have the service manual / diagrams for this system? I need a copy asap.
getting error 62 (motion) when doing a LMAO using either bucky

Log inor Register

to rate and post a comment

(2)

Eric Hazen

re: re: Lorad MIV - System Error 33 & Bucky Error 51

March 25, 2015 03:08

If I'm only getting the error 33 would that also be related to the motors or would the IR pcb be more likely? Thanks for any help, eric_hazen@ssmhc.com

Log inor Register

to rate and post a comment

(130)
(32)

FranciscoG

re: Lorad MIV - System Error 33 & Bucky Error 51

April 26, 2014 11:43

Thanks JeffS and James McGinty, the grid drive motor in the HTC Bucky needed to be replaced and problem solved! I appreciate your help a lot!

Log inor Register

to rate and post a comment

(66)
(12)

gerardo millena

re: Lorad MIV - System Error 33 & Bucky Error 51

April 28, 2014 08:56

Ok so it's the grid motor, that's great. I have a trick or bandage awaitng motor replacement. Try remove the LAN cable going to the Bucky grid. During power-up the system does not see that this is a part of the system and should permit exposure I think......

1 person liked this
Log inor Register

to rate and post a comment

Sentient Lasers the SOLUTIONS Company Service, Sales, Buy, Warranty & Lease

The only company with Blue Dot Certified pre-owned lasers & energy based devices. Ask about 1-year warranties on top OEM’s Cutera, BTL, Candela, Lumenis, Palomar, Alma, Syneron & Cynosure! (855) 819-3781 r.haller@sentientlasers.com www.SentientLasers.com

DOTmed Certified DOTmed 100
advertisement

 

RSTI Newsletter

 

one source

 

Hitachi