Fuji Fcr Xl-2 Carbon Error 11389 - Servicetechniker Forum

Fuji FCR XL-2 Carbon: Error '11389'

Tags: PACS / DR / CR / Digital

Posted by an Anonymous User

Fuji FCR XL-2 Carbon: Error '11389'

March 06, 2017 09:08

I have a Fuji FCR XL Carbon that give error '11389' anytime I read a cassette. I can tried multiple time but it still doesn't read the cassete.

Does anyone know how to fix this?

Regards.

Log inor Register

to rate and post a comment

(2677)
(605)
(17)

Lenny Place

re: Fuji FCR XL-2 Carbon: Error '11389'

March 07, 2017 10:15

11389 IP suction error (before IP
reading)
The IP suction pump (PA1) turned ON during routine processing to perform an IP suction operation.
However, it was found that the IP suction sensor (SA5) was CLOSED (the IP was not vacuum-retained).
Although retries were performed, the maximum retry count was exceeded

Log inor Register

to rate and post a comment

(1)

LadyBiomed

re: re: Fuji FCR XL-2 Carbon: Error '11389'

June 11, 2019 08:51

Hello I am receiving the same error. By chance do you have a service manual or could you tell me the location of the SA5 sensor?

Log inor Register

to rate and post a comment

(2677)
(605)
(17)

Lenny Place

re: re: re: Fuji FCR XL-2 Carbon: Error '11389'

June 17, 2019 05:12

if you send an email address i will send you the service manual.
the SA5 sensor is accessed on the backside top section. near the vacuum pump

Log inor Register

to rate and post a comment

(3)

Carlos Jiminian

re: re: re: re: Fuji FCR XL-2 Carbon: Error '11389'

July 11, 2019 09:56

hi Lenny!
you can send me service manual?
thanks You

Log inor Register

to rate and post a comment

(1)

FrankB

re: Fuji FCR XL-2 Carbon: Error '11389'

March 07, 2017 10:24

That error would indicate a problem with the IP Suction Sensor SA5. There could be dust causing it to not function properly or it may need to be replaced.

Log inor Register

to rate and post a comment

THE (LEADER) IN MEDICAL IMAGING TECHNOLOGY SINCE 1982. SALES-SERVICE-REPAIR

Special-Pricing Available on Medical Displays, Patient Monitors, Recorders, Printers, Media, Ultrasound Machines, and Cameras. This includes Top Brands such as SONY, BARCO, NDS, NEC, LG, EDAN, EIZO, ELO, FSN, PANASONIC, MITSUBISHI, OLYMPUS, & WIDE.

advertisement

 

RSTI Newsletter

 

one source

 

 

Hitachi