Drgem Gxr32 - Servicetechniker Forum
advertisement
Aktueller Standort:
>
>
> Drgem GXR32

Drgem GXR32

(7)

Viramani R Ranganathan

Drgem GXR32

August 09, 2019 10:01

Anyone familiar with GXR32 Drgem Generator , 32kw

Recently I troubleshoot generator and found Main Earth Electric Circuit Breaker ELCB closed when Turn ON generator.

I checked and found IGBT GX power board J1 connected cause main ELCB closed.

I disconnected J1 connection from GX Power generator able turn On but show error 32.

Further test ELCB still closed when press main contactor inside generator even generator insulator switch in off ( 0Vac) condition.

Please help

Log inor Register

to rate and post a comment

(6)
(1)

Geoff Lodge

re: Drgem GXR32

August 11, 2019 02:24

The only issue I have ever come across with these units is the electrolytic capacitors on the IGBT gate board. Closely inspect for swollen cases and replace if required. I believe they are 50V 330uF electrolytic capacitors.

Log inor Register

to rate and post a comment

(7)

Viramani R Ranganathan

re: re: Drgem GXR32

August 13, 2019 10:42

Can I know what kind issue you are face it?

Log inor Register

to rate and post a comment

Servicing GE/Siemens Nuclear Medicine equipment with OEM trained engineers

Numed, a well established company in business since 1975 provides a wide range of service options including time & material service, PM only contracts, full service contracts, labor only contracts & system relocation. Call 800 96 Numed for more info.

 

advertisement

 

advertisement

 

advertisement

 

advertisement
advertisement