Can We Install Dura Tube For Siemens Sensation 16 Ct Instead Of - Servicetechniker Forum
advertisement

Can we install Dura tube for Siemens sensation 16 CT instead of

Tags: CT, SIEMENS

(6)

(2)

Walid Abu Farha

Can we install Dura tube for Siemens sensation 16 CT instead of

October 02, 2019 10:12

Can we install Dura tube for Siemens sensation 16 CT instead of Straton Tube

Log inor Register

to rate and post a comment

(27)
(4)

Eric Hauk

re: Can we install Dura tube for Siemens sensation 16 CT instead of

October 03, 2019 10:11

You can but takes some modifications to the Cbox i believe. It’s been a long time since I tackled that project

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: Can we install Dura tube for Siemens sensation 16 CT instead of

October 03, 2019 10:11

Yes, but you need to change the HV system,and softwrae. big job

Log inor Register

to rate and post a comment

(296)
(52)

Gabriel Cool

re: Can we install Dura tube for Siemens sensation 16 CT instead of

October 03, 2019 10:17

No

Log inor Register

to rate and post a comment

(1)

HAssadi

re: Can we install Dura tube for Siemens sensation 16 CT instead of

October 03, 2019 03:24

Yes you can but it requires alot of work. The gantry weights must be swapped out with a set from an akron q or dunlee tube.

Log inor Register

to rate and post a comment

(296)
(52)

Gabriel Cool

re: re: Can we install Dura tube for Siemens sensation 16 CT instead of

October 04, 2019 10:11

No he can't. It's will be easy to change totally gentry. He need ti change colimatir, ebox, generator, DOM. And how he can do this?

Log inor Register

to rate and post a comment

Nationwide Service & Support 1-866-NUC-MED1

OEM trained nationwide service for NM, PET, PET/CT systems. Relocation, installation, parts, technical and clinical support. Contact us at info@nis-mit.com

advertisement

 

advertisement

 

advertisement

 

advertisement