Hello, we are a small non profit in Minnesota doing work in Haiti. We have a siemens mobilett plus hp x-ray unit in a pediatric medical center in Port Salut, Haiti. We are looking for a service/operations manual for the unit. Please respond to sgrundhoffer@notimeforpoverty.org. Thanks so much!
I am looking for Siemens Mobilette service and operators manual for International Children's Heart Foundation for our Haiti program in Jimani. I would appreciate any help !
Roy Morris,CBET
Roy@babyheart.org
PLease Help. I am looking for operator and service manuals for
1. Siemens Mobilette Plus Mobile X Ray
2. Cuda Gemini Light Source Headlight Head Light Fiber Optic System on Stand Model M2-150T. This is needed for The Anamayah Memorial Hospital in Guyana South America. I would appreciate any help.
Thanks
Ian Anamayah
iananamayah@yahoo.com
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
March 12, 2015 11:29
HI good day,Dear Morris, I need a Service & operational manual for the Siemens Mobilett PLUS.Could you please send me a copy at my e mail biodiem@gmailcom. Thanks
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 25, 2016 02:29
Good day sir, I have a problem (err23) with a Mobile Plus, would you be so kind to send me a copy of the service manual?
my e-mail is pedrojesus88@hotmail.com
re: re: re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 29, 2016 10:07
Very possible a tube arc
Err 23 During exposure The HT inverter has
been interrupted
several times, e.g.
overcurrents.
When the HT inverter has been interrupted
more than 3 times, the exposure
will be interrupted.
⢠Faulty inverter/inverter control
⢠Check the inverter current, refer to
the section "Error diagnosis HT
and single tank". Refer also to
"P12 - Inverter and single tank test"
and "P13 - Inverter test".
⢠The X-ray tube may have arced
during the exposure. If the first
exposure after a long interval is an
exposure with high kV values, tube
arcs can occur.
The unit has to be started with
several exposures at high mAs
values. The kV values selected
should be from 40 kV increased to
133 kV.
Refer also to the section "Error
diagnosis HT and single tank".
Estimated Roy Morris, have a wonderful day and success in your daily work . I need a Service & operational manual for the Siemens Mobilett PLUS.Could you please send me a copy to my e-mail ,bhdez1966@gmail.com . thanks you for your help. bests regards
ask friends x-ray diagrams siemens Mobilett hp, manual I have, I have problems with error 17, followed the directions in the manual and not solved it, I need the capacitor charging circuit; BUZ330 possible failure.
thank you
ask friends x-ray diagrams siemens Mobilett hp, manual I have, I have problems with error 17, followed the directions in the manual and not solved it, I need the capacitor charging circuit; BUZ330 possible failure.
thank you
I am also in need of this manual for International Children's Heart Foundation. We are starting our first Medical Mission to Haiti on April 20, 2013. I sent them a Siemens Mobilette four months ago and need a manual. Thanks. Roy Morris,CBET
Sue Grundhoffer
Siemens Mobilett Plus HP - manual
January 09, 2013 11:29
Hello, we are a small non profit in Minnesota doing work in Haiti. We have a siemens mobilett plus hp x-ray unit in a pediatric medical center in Port Salut, Haiti. We are looking for a service/operations manual for the unit. Please respond to sgrundhoffer@notimeforpoverty.org. Thanks so much!
to rate and post a comment
Ralph Frizzle
re: Siemens Mobilett Plus HP - manual
January 09, 2013 05:13
please contact our office technical dept
to rate and post a comment
(8)
Roy Morris
re: re: Siemens Mobilett Plus HP - manual
January 09, 2013 08:39
I am looking for Siemens Mobilette service and operators manual for International Children's Heart Foundation for our Haiti program in Jimani. I would appreciate any help !
Roy Morris,CBET
Roy@babyheart.org
to rate and post a comment
Lenny Place
re: re: re: Siemens Mobilett Plus HP - manual
January 10, 2013 05:29
you will get a email from sendthisfile
to rate and post a comment
Ian Anamayah
re: re: re: re: Siemens Mobilett Plus HP - manual
September 14, 2013 09:37
PLease Help. I am looking for operator and service manuals for
1. Siemens Mobilette Plus Mobile X Ray
2. Cuda Gemini Light Source Headlight Head Light Fiber Optic System on Stand Model M2-150T. This is needed for The Anamayah Memorial Hospital in Guyana South America. I would appreciate any help.
Thanks
Ian Anamayah
iananamayah@yahoo.com
to rate and post a comment
(8)
Roy Morris
re: re: re: re: re: Siemens Mobilett Plus HP - manual
February 14, 2014 10:06
I have a copy but need your email to send it to you. Email me at roy@babyheart.org
to rate and post a comment
(1)
Tunji Amosu
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
February 22, 2014 08:23
Dear Roy, have message you on your E-mail, you can kindly send it to us through our E-mail : Boltsmed@yahoo.com
to rate and post a comment
(8)
Ernest ONUIGBO
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
February 26, 2014 08:22
Dear Morris, I need a Service & operational manual for the Siemens Mobilett PLUS.Could you please send me a copy via nestob2002@yahoo.com. Thanks
to rate and post a comment
(23)
Tony Smith
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
May 21, 2014 06:09
tsmith@udrconversion.com
to rate and post a comment
Engineer
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
December 22, 2014 10:11
can you send to me too , Eng_mohd81@yahoo.com
to rate and post a comment
GARCIA ANTONIO
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
March 12, 2015 11:29
HI good day,Dear Morris, I need a Service & operational manual for the Siemens Mobilett PLUS.Could you please send me a copy at my e mail biodiem@gmailcom. Thanks
to rate and post a comment
Patrick Paget
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
April 16, 2015 01:38
here also please patrickpaget@gmail.com
to rate and post a comment
(2)
(1)
Faheem Saleemi
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
October 05, 2015 07:47
Mr.Roy Morris, Please send me the copy of siemens mobilett plus hp at saleemi4u@gmail.com best regards. Faheem Saleemi
to rate and post a comment
(33)
FaheemSaleemi
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
October 06, 2015 08:26
I am thank-full to u for your cooperation.
to rate and post a comment
Pedro Jesus Arenas Calderon
re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 25, 2016 02:29
Good day sir, I have a problem (err23) with a Mobile Plus, would you be so kind to send me a copy of the service manual?
my e-mail is pedrojesus88@hotmail.com
to rate and post a comment
DMEDS
re: re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 27, 2016 05:47
Dear,
I have the manuals for this system if you need it contact me,
âRegardsâ
magmesme@gmail.com
to rate and post a comment
DMEDS
re: re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 27, 2016 05:47
Err 23 During exposure, the main inverter was interrupted >3 times
to rate and post a comment
DMEDS
re: re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 27, 2016 05:47
Err 23 During exposure The HT inverter has
been interrupted
several times, e.g.
overcurrents.
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 29, 2016 10:07
Very possible a tube arc
Err 23 During exposure The HT inverter has
been interrupted
several times, e.g.
overcurrents.
When the HT inverter has been interrupted
more than 3 times, the exposure
will be interrupted.
⢠Faulty inverter/inverter control
⢠Check the inverter current, refer to
the section "Error diagnosis HT
and single tank". Refer also to
"P12 - Inverter and single tank test"
and "P13 - Inverter test".
⢠The X-ray tube may have arced
during the exposure. If the first
exposure after a long interval is an
exposure with high kV values, tube
arcs can occur.
The unit has to be started with
several exposures at high mAs
values. The kV values selected
should be from 40 kV increased to
133 kV.
Refer also to the section "Error
diagnosis HT and single tank".
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: Siemens Mobilett Plus HP - manual
November 29, 2016 10:08
ok
to rate and post a comment
(5)
Emanuel Batista
re: re: re: re: re: Siemens Mobilett Plus HP - manual
March 09, 2017 10:39
hello, you can send me this service manual and operator manual please.
to rate and post a comment
Rene Serna
re: re: re: re: Siemens Mobilett Plus HP - manual
March 11, 2015 04:06
Hi Lenny Place:
Could you be so gentile and send a copy of those manuals to me please.
I will be waiting for your kindly reply.
Many thanks and best regards.
Rene Serna
to rate and post a comment
(1)
lei wang
re: re: re: re: Siemens Mobilett Plus HP - manual
September 09, 2018 11:19
DEAR MR Lenny Place
I NEED THE MANUAL, YOU MAY SEND ME A COPY weida-02@163.com I WOULD APRECCIATE THE HELP.
to rate and post a comment
(1)
bioengineer
re: re: re: Siemens Mobilett Plus HP - manual
May 04, 2018 10:15
Estimated Roy Morris, have a wonderful day and success in your daily work . I need a Service & operational manual for the Siemens Mobilett PLUS.Could you please send me a copy to my e-mail ,bhdez1966@gmail.com . thanks you for your help. bests regards
to rate and post a comment
(1)
JOSEPH IGASHI
re: re: Siemens Mobilett Plus HP - manual
February 14, 2014 09:00
Please I need operational manual for Siemens mobilett plus, kindly help me. Shepherdmedicals2010@hotmail.com
to rate and post a comment
(8)
Roy Morris
re: re: re: Siemens Mobilett Plus HP - manual
February 14, 2014 10:06
Send me your email and I will help you. Roy@babyheart.org
to rate and post a comment
(5)
kco
re: re: Siemens Mobilett Plus HP - manual
December 18, 2014 12:26
ask friends x-ray diagrams siemens Mobilett hp, manual I have, I have problems with error 17, followed the directions in the manual and not solved it, I need the capacitor charging circuit; BUZ330 possible failure.
thank you
to rate and post a comment
(5)
kco
re: re: re: Siemens Mobilett Plus HP - manual
December 19, 2014 10:17
ask friends x-ray diagrams siemens Mobilett hp, manual I have, I have problems with error 17, followed the directions in the manual and not solved it, I need the capacitor charging circuit; BUZ330 possible failure.
thank you
my mail floresr_kco@hotmail.com
to rate and post a comment
Lenny Place
re: Siemens Mobilett Plus HP - manual
January 09, 2013 06:07
you will get a email from sendthisfile
to rate and post a comment
Patrick Paget
re: re: Siemens Mobilett Plus HP - manual
April 16, 2015 01:38
please send to me also at patrickpaget@gmail.com
to rate and post a comment
(8)
Roy Morris
re: Siemens Mobilett Plus HP - manual
January 09, 2013 08:39
I am also in need of this manual for International Children's Heart Foundation. We are starting our first Medical Mission to Haiti on April 20, 2013. I sent them a Siemens Mobilette four months ago and need a manual. Thanks. Roy Morris,CBET
to rate and post a comment
(8)
Roy Morris
re: re: Siemens Mobilett Plus HP - manual
January 11, 2013 08:25
Thanks for the manual!
to rate and post a comment
Patrick Paget
re: re: re: Siemens Mobilett Plus HP - manual
April 16, 2015 01:38
patrickpaget@gmail.com please send a copy
to rate and post a comment
ARMANDO BARAJAS
re: re: re: Siemens Mobilett Plus HP - manual
August 25, 2018 11:02
DEAR MR ROY MORRIS
I NEED THE MANUAL, YOU MAY SEND ME A COPY acbbara@gmail.com I WOULD APRECCIATE THE HELP.
to rate and post a comment
Lenny Place
re: re: re: re: Siemens Mobilett Plus HP - manual
August 26, 2018 11:29
I am sending it to you using myairbridge an online site
to rate and post a comment
(1)
lei wang
re: re: re: re: re: Siemens Mobilett Plus HP - manual
September 09, 2018 11:19
DEAR MR Lenny Place
I NEED THE MANUAL, YOU MAY SEND ME A COPY weida-02@163.com I WOULD APRECCIATE THE HELP.
to rate and post a comment
(1)
lei wang
re: re: re: Siemens Mobilett Plus HP - manual
September 09, 2018 11:19
DEAR MR ROY MORRIS
I NEED THE MANUAL, YOU MAY SEND ME A COPY weida-02@163.com I WOULD APRECCIATE THE HELP.
to rate and post a comment
mohamed edris
re: re: re: Siemens Mobilett Plus HP - manual
October 01, 2018 08:57
I need user manual
to rate and post a comment