Emmanuel Quaye
SEDECAL SHF 330 communication error E74
September 11, 2020 09:38
SEDECAL SHF 330. I have communication error E74 After exiting service Mode.
to rate and post a comment
(1)
(1)
Howard Karp
re: SEDECAL SHF 330 communication error E74
September 14, 2020 10:31
I looked up Error 74: excessive current in main winding braking at 3300 rpm-CHECK CR8, 9, 10, 11 in Interface DRAC-HF. Also check Rotor wiring for shorts to ground.