Philips BV libra 1.2 Error 341

July 12, 2019 11:01

Hi.
I have a Philips Bv libra R. 1.2 with the error 341 I have the service manual SMCM for BV family 1.2 and the description of the error is
"X-ray generator not available Switch system off/on. Call service if problem is persistent RA generator board SHA4 Generator ready signal is inactive. Check generator board. Check generator unit"
I don’t have the Schematics to follow the ready signal, from the Generator, any one had this problem before or can send me the Schematics.
email me roberto.vaca@ingevidasrl.com
Thanks

Log inor Register

to rate and post a comment

(34)
(4)
(2)

Thameen Ansari

re: Philips BV libra 1.2 Error 341

July 14, 2019 03:02

Connect D connector to SHA4 contact X2 point 19. Is 15 s after system start the result between 10 and 15 V. If YES, Replace SHA4.
If No LED of the PS lights up. If No Check PS.
If Yes - Replace generator.

1 person liked this