Hello folks, i replace memory`s SRAM battery do factory reset but i get this error message on the top up of the screen Select config. , But the SPO2 nellcor doesn't work and the finger probe is lighting up.

I'd appreciate so much if you could tell me how to solved the problem.

Log inor Register

to rate and post a comment

(6)

Carlos Velasco

re: Need help with Datex Ohmeda monitor Please.

November 23, 2021 04:38

If the manufacturing time of the Time Keepper is greater than 2 years, what happens is that the internal chip Nickel Cadmium batteries have not been charged in a long time and it may take time to recover. Leave the equipment on and wait the necessary time for them to retain charge and to retain the configuration information. I have waited up to a week for the problem to be resolved. The best thing is to have fresh chips and so when they are changed immediately the problem is corrected

Log inor Register

to rate and post a comment

(9)
(1)

Borislav Ivanov

re: re: Need help with Datex Ohmeda monitor Please.

November 24, 2021 10:40

Can you explain me of this Nickel Cadmium batteries please? so for this i should leave the device ON and wait .

Log inor Register

to rate and post a comment

(6)

Carlos Velasco

re: re: re: Need help with Datex Ohmeda monitor Please.

November 29, 2021 10:10

If, like any rechargeable battery, if it does not charge and discharge cycles, the memory effect is generated and they lose storage capacity. As they are for a long time without doing these cycles, they are downloaded and it has to take a while to regain their functionality. If the manufacturing date is far from the current time, they may not even be recovered. I have waited up to a week, then I better replace it with another

Log inor Register

to rate and post a comment

(409)
(82)
(3)

juan pablo amado

re: Need help with Datex Ohmeda monitor Please.

November 23, 2021 04:44

Datex sensor SpO2 is not compatible with Nellcor.

Log inor Register

to rate and post a comment

(1)
(1)
(1)

Mike Balakonis

re: Need help with Datex Ohmeda monitor Please.

November 23, 2021 04:44

From what I remember. You must discharge the patient after changing out the SRAM battery and then do a reboot.

1 person liked this
Log inor Register

to rate and post a comment

(6)

Carlos Velasco

re: re: Need help with Datex Ohmeda monitor Please.

November 24, 2021 10:39

You change the battery, set the language, time and date. You turn it off and confirm that it retains the data. From there you can reconfigure all the equipment. If you turn it off and retain the configuration, it is already there and you can register it, check the physiological parameters with the patient simulator.

Log inor Register

to rate and post a comment

(6)

Carlos Velasco

re: re: Need help with Datex Ohmeda monitor Please.

November 24, 2021 10:40

What we do in the company is to place a chip and leave the equipment on, after a few hours we set the correct time and date and leave a few more minutes and turn it off. We turn it on after about 5 minutes and if it retains the time and date, we proceed to configure the language. network frequency, etc. It is left for a few minutes and it is paid, you wait 5 minutes and turn it on again. If you retain the chosen data the monitor is ready

Log inor Register

to rate and post a comment

(9)
(1)

Borislav Ivanov

re: re: Need help with Datex Ohmeda monitor Please.

November 24, 2021 10:41

How can i discharge the patient ?