konica reader 110 regius image process issues

Tags: PACS / DR / CR / Digital, KONICA MINOLTA

(43)
(8)

ATLtd

konica reader 110 regius image process issues

November 22, 2021 11:18

hi
Regius 110 reader showing the image without any object or body parts in the x-ray processed image . image is grey . the s value showing 1500. this confirm the plate received the x-ray. no error from the reader or console. please let me know what is the cause for this image issue..

many thanks
ATL

Log inor Register

to rate and post a comment

(303)
(47)

CRAIG WAMBURA

re: konica reader 110 regius image process issues

November 23, 2021 04:44

1. Set the S-Value to the appropriate value
2. Set the ROI to the correct position
3. Check photomultiplier switch and ensure its ON
4. Could be a fault in LMC. Replace
5. Could be fault in photomultiplier. Replace its filter assay

Log inor Register

to rate and post a comment

(43)
(8)

ATLtd

re: re: konica reader 110 regius image process issues

November 29, 2021 10:11

thank you. All above checked . Noticed S number is above 1400 for problem image and s value for normal images are 2 . Optics checked. Recalibrated the reader. CR Cassettes are OK. Still intermittent image issue.. Please let me know what is the next step..

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Cyclic Materials - Sell us your permanent magnet MRIs! WE BUY WARM MAGNETS FOR CASH

Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: david.aguiar@cyclicmaterials.earth

stats
advertisement

Upcoming Auction Events

NorthWest Supply
TIMED
Apr 08, 2024 (99 lots)