Kodak 8000 Pano Carestream Error/Calibaration

Tags: Dental, KODAK CARESTREAM

Posted by an Anonymous User

Kodak 8000 Pano Carestream Error/Calibaration

April 20, 2020 10:12

I have Kodak 8000 Carestream Pano, whenever turned on always shows different errors- sometimes shows E15, E53, or cooling tube although we do not use radiation.
Please help. Thank you

Log inor Register

to rate and post a comment

(329)
(220)
(4)

Faustino Pereira

re: Kodak 8000 Pano Carestream Error/Calibaration

April 21, 2020 10:16

Change the RTC (Real Time Clock on CPU board. It can be a
M48T86 thick black IC or an yellow battery inserted on an IC.
After replacement you get err 58, then cacel the error and make a full pan. Make sure you insert the components in the right direction. Thanks for reading
.

1 person liked this
Log inor Register

to rate and post a comment

(2)

Syed Naseer

re: re: Kodak 8000 Pano Carestream Error/Calibaration

April 22, 2020 10:27

I did change the battery as it was not able to take any pano. Since then it’s giving the different errors mentioned above.
It has yellow battery. I do not know may be new battery is also expired( bought from amazon) . Do you have any suggestion, from where I can buy one?
Thank you.

Log inor Register

to rate and post a comment

Evgeniy Ильинский

re: re: Kodak 8000 Pano Carestream Error/Calibaration

June 02, 2020 11:13

Hi,
I also have the E53 problem, can you send me the manual to fix it?
thx

Log inor Register

to rate and post a comment

(329)
(220)
(4)

Faustino Pereira

re: re: re: Kodak 8000 Pano Carestream Error/Calibaration

June 03, 2020 09:42

Hi, The solution has been sent by email to you.
Thanks for reading,

1 person liked this
Log inor Register

to rate and post a comment

(2)

Donnell Greene

re: re: re: re: Kodak 8000 Pano Carestream Error/Calibaration

June 28, 2022 10:17

Please send me the solution also.
Had E53 and replaced battery.
When attempt to go in menu for calibration got E51.
After a week E51 went away.
When attempt to go in menu for calibration again got E51.
Any suggestions.