Medilab Global provides high-quality pre-owned imaging equipment, parts, and service to the healthcare industry worldwide. Our number 1 position is to provide excellent customer service.. 305-234-0084 | Email: info@medilabglobal.com | Hablamos Español
(2)
MONA
I NEED SERVICE MANUAL Polydoros LX 50
January 29, 2014 08:24
Dear All
I HAVE ERROR 606
Looking for a service manual for Routine X-Ray, Polydoros LX 50 SIEMENS.
Regards
to rate and post a comment
Lenny Place
re: I NEED SERVICE MANUAL Polydoros LX 50
January 29, 2014 06:13
I can send it to you email address please
606 Meaning:
RANGE Error
Description:
The signals RANGE0 ”OK” and RANGE1 ”OK” monitor the switchover of the tube current measuring ranges on D220. The corresponding relays are monitored via the RANGE 0, and RANGE 1 signals.
Possible causes and action:
Check function at SS OFF on D220. Test points: D220.X19.A1, A2
(wiring diagram X2206-18 and 22 / X2169-18 and 22):
1. With ZB ON: measuring range 200mA;
"RANGE 0" X19.A1 = 0V -> "RANGE 0" ”OK” X19.A4 = 24V (±15%)
"RANGE 1" without significance
2. With normal FL ON: measuring range 1mA;
"RANGE 0" X19.A1 = 24 V (±15%)
"RANGE 0" ”OK” X19.A4 = 0V
RANGE 1" X19.A2 = 24V (±15%) -> "RANGE 1" ”OK” X19.A3 = 0V
If the "RANGE 0" and "RANGE 1" signals do not agree with the operating mode, there is a fault in the cable X1 to D100 or on D100.
If the signals "RANGE 0"_OK and "RANGE 1"_OK are wrong with correct actuation, D220 must be replaced.
XCS ©2004 by Siemens AG Med AXPLM-T SC
to rate and post a comment
(2)
MONA
re: re: I NEED SERVICE MANUAL Polydoros LX 50
January 31, 2014 04:25
Dear Lenny Place
Thank you for your guidance.
Could you please send for me service manual ?
my email : monataqavi@gmail.com
thank you & Best regards
Mona
to rate and post a comment
(36)
Luis Fernando Garcia Sanchez
re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
December 02, 2015 09:43
Could you send me?, service manual polydoros lx. Email luisfernandogarcias@yahoo.com regards
to rate and post a comment
DMEDS
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
December 09, 2015 08:50
i have all manuals for this system
PolyLX_Adjustment
PolyLX_Troubleshooting
PolyLX_Wiring and schematic diagram
Replacement of Parts
contact me
magmesme@gmail.com
regards
to rate and post a comment
(1)
Alejandro Orihuela
re: re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
February 06, 2019 08:48
regards!! I need the manual of polydoros lx 50, I get error 805 and 777 thanks
Please send me the service manual please
alejandro_orihuela@hotmail.com
to rate and post a comment
Lenny Place
re: re: re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
February 06, 2019 09:10
Look for an email from a site called myairbridge
to rate and post a comment
(51)
(1)
Amos Dantene
re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
November 07, 2018 08:21
Please I need the service manual. My email address is: adantene@gmail.com
to rate and post a comment
Lenny Place
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
November 08, 2018 09:55
Look for an email from a site called myairbridge
to rate and post a comment
Rama Kant
re: re: I NEED SERVICE MANUAL Polydoros LX 50
June 12, 2015 03:44
Dear Lenny Place,
I shall be thankful for sending me servicr manual of Polydoros LX 50 at my mail reconsystems@yahoo.com
Regards
Rama Kant
to rate and post a comment
DMEDS
re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
December 09, 2015 08:50
i have all manuals for this system
PolyLX_Adjustment
PolyLX_Troubleshooting
PolyLX_Wiring and schematic diagram
Replacement of Parts
contact me
magmesme@gmail.com
regards
to rate and post a comment
(2)
nos
re: re: I NEED SERVICE MANUAL Polydoros LX 50
April 17, 2018 10:47
what does err 119/080 for generator POLYDOROS LX 50 mean
to rate and post a comment
Lenny Place
re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
April 17, 2018 05:58
Error 119
Designation:
Heating circuit timeout
Description:
Communication between the D220 heating circuit and the D100 master takes place over the serial interface (MPS, D100.X1.28, D220.X1.28)
If an interrupt of communication between the D100 and D220 takes place, error 119 is sent out by the D100.
Possible Causes of Error and Remedies:
· Serial interface interrupted, check â
· D100, D220 defective, replace â
· Arcing tube
· H1 high voltage transformer
· Check how the cables are laid (shielding)
· Check the ground wires, all ground wire connections provided by the manufacturer must be connected.
· Check the firmware version.
· Check the 5V power supply on the SNT, D160 and D220.
- Measure at the SNT X2.6 (5V) switch power supply versus X2.3 (0V), set approx. 5.1V (max. 5.2V).
- On the D160. Measure X31 (5V) versus X28, the voltage drop versus the SNT should be less than 20mV.
- Measure on D220.X18.A1 (5V) versus X15, the voltage drop versus the SNT should be less than 150mV.
- If the voltage drop on the D220 is more than 150mV, check the following plug-in connections: SNT.X2, D160.X11, D160.X13, D100.X13, D100.X1, D220.X1.
- If the X1 ribbon cable between D100 and D2209 is causing the voltage drop, replace it with a new one with Part No. 7463677.
· Observe the generator Overview Wiring Diagram for lamps and LEDs in the power cabinet.
· Ignore error 119 if the red LED (D220) is blinking and no firmware is loaded; perform a â download.
to rate and post a comment
Reinaldo Costa
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
May 15, 2018 12:02
Please, I NEED service manual Polydoros LX 50. how much? My e-mail: reinaldo@reitecequipamentos.com.br
to rate and post a comment
Reinaldo Costa
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
May 15, 2018 12:02
Please, I NEED service manual Polydoros LX 50. how much? My e-mail: reinaldo@reitecequipamentos.com.br
to rate and post a comment
Lenny Place
re: re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
May 16, 2018 09:20
I am sending to you
to rate and post a comment
(8)
Peter Russell
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
January 10, 2019 09:35
I should have said p.russell@medimaging.co.uk
to rate and post a comment
houssein ibraheem
re: re: I NEED SERVICE MANUAL Polydoros LX 50
April 23, 2018 10:56
hello
i have error 865/016
could you please advice me,
please send me the service manual please
bme.houssein@gmail.com
to rate and post a comment
Lenny Place
re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
April 23, 2018 04:09
what system are you working on this is not a generator error code? normally a APID 16 deals with a videomed or imaging component and 865 is not showing on my list.
to rate and post a comment
houssein ibraheem
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
April 30, 2018 09:42
the table model is sireskop cx
now i facing a new error 007/035 sireskop cx
could you please send me the service manual?
my email is bme.houssein@gmail.com
Thanx
BR
to rate and post a comment
houssein ibraheem
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
May 28, 2018 10:36
good day,
these error appears on the screen
242/032
246/032
248/032
any advice please
the system is sireskope cx with polydoros lx50
to rate and post a comment
(8)
Peter Russell
re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50
January 10, 2019 09:35
Hi Lenny could you send me the manual please. I have error 777 on a Tosrad with a lite generator
to rate and post a comment
sai KUMAR
re: re: I NEED SERVICE MANUAL Polydoros LX 50
March 24, 2019 04:00
dear sir
i have error looking for service manual schematic polydoros lx 50 plz sir
to rate and post a comment
sai KUMAR
re: re: I NEED SERVICE MANUAL Polydoros LX 50
March 24, 2019 04:00
Hi sir my name is Saikumar sir I want polydoros Lx 50 schematic diagram sir please help me
to rate and post a comment
Mohammed Abdullah
re: I NEED SERVICE MANUAL Polydoros LX 50
January 29, 2014 06:14
i have all manuals for this system
PolyLX_Adjustment
PolyLX_Troubleshooting
PolyLX_Wiring and schematic diagram
Replacement of Parts
contact me
magmesme@gmail.com
regards
to rate and post a comment
Reinaldo Costa
re: I NEED SERVICE MANUAL Polydoros LX 50
May 15, 2018 12:02
Please, I NEED service manual Polydoros LX 50. how much? My e-mail: reinaldo@reitecequipamentos.com.br
to rate and post a comment