Hi friends!
I have a problem with 2 GE ups on angiographys. After two years beginning lamp flashes "service check".
In the service manual for angiography Innova 3100 is written, that need have to do service work, but what i must do is not written.
What should I do to remove this flashing?
Thanks
This sounds similar to an ultrasound system when the backup battery is low on voltage. Check the backup batteries for the proper voltages and make sure the batteries aren't swollen and leaking, because they can crack.
is this the GE SG series UPS? the service light can be on if service is needed (whihc model do you have I) or if Q1 is open, indicating that the inverter is in service mode and not supplying the load.
go to this site and download the UPSMAN software you will be able to retrieve the complete event logs
http://www.generex.de/index.php?option=com_content&task=view&id=62&Itemid=102
Hello dear, I'm from Venezuela and I need help to take the alarm off the service to a GE SG series 0 80KVA UPS, if you know how to do it I would appreciate it if you could share that information with me, since the maintenance was done to the team but the Alarm does not disappear. Thank you
Hello dear, I'm from Venezuela and I need help to take the alarm off the service to a GE SG series 0 80KVA UPS, if you know how to do it I would appreciate it if you could share that information with me, since the maintenance was done to the team but the Alarm does not disappear. Thank you
Hi Lenny, Do you happen to have any documentation to help me troubleshoot GE SG Series 20kVA UPS. I have all the manuals included on GE Innova 2100/3100/4100IQ SM (under OEM section) but its limited. We have 4 systems which all appear to be error free (all have service light blinking ofcourse) but one of the units does not switch the load to battery (even though the inverter is on and load connected with fresh batteries instaled). UPS just shuts off when removing power. Have checked the wiring and all messages and events, see nothing out of the ordinary. Need documentation else I have to wait 2 weeks for UPS company. We are well versed with UPS systems (and can fully service other OEMS), but GE has extremely limited documentation Thanks in advance for any help you can provide. matt@tranzrad.com
Hi Dear. I have a problem with pm. I already replace the capacitors and fans, but wen te ups starting i hav a massage replacw it. Can you help me to tell wich values need to change. And how i can gererate the service key because the service password is don't matching. Thanks advice.
(27)
Aleksander Proof
GE UPS service check
December 30, 2013 10:24
Hi friends!
I have a problem with 2 GE ups on angiographys. After two years beginning lamp flashes "service check".
In the service manual for angiography Innova 3100 is written, that need have to do service work, but what i must do is not written.
What should I do to remove this flashing?
Thanks
to rate and post a comment
(21)
(32)
Zach Ruder
re: GE UPS service check
December 31, 2013 06:04
This sounds similar to an ultrasound system when the backup battery is low on voltage. Check the backup batteries for the proper voltages and make sure the batteries aren't swollen and leaking, because they can crack.
to rate and post a comment
Lenny Place
re: GE UPS service check
December 31, 2013 06:04
is this the GE SG series UPS? the service light can be on if service is needed (whihc model do you have I) or if Q1 is open, indicating that the inverter is in service mode and not supplying the load.
to rate and post a comment
(27)
Aleksander Proof
re: re: GE UPS service check
January 03, 2014 07:20
Yes its GE SG series. And its need service. What i must do on service?
to rate and post a comment
Lenny Place
re: re: re: GE UPS service check
January 03, 2014 09:28
can you retrieve the error code the system is getting from there I can tell you what to check
to rate and post a comment
Lenin Jose
re: re: re: re: GE UPS service check
November 21, 2016 02:57
Hello Friend Where can I get the manual of service of the UPS SG series 300, I have a problem with the inverter.
to rate and post a comment
Lenny Place
re: re: re: GE UPS service check
January 03, 2014 09:29
go to this site and download the UPSMAN software you will be able to retrieve the complete event logs
http://www.generex.de/index.php?option=com_content&task=view&id=62&Itemid=102
to rate and post a comment
(3)
Edward Larkin
re: re: re: re: GE UPS service check
November 03, 2017 04:01
Do you know Which Generex (GE) version? 12 or 23?
to rate and post a comment
(3)
Edward Larkin
re: re: re: re: GE UPS service check
November 03, 2017 04:01
Service Manual available?
to rate and post a comment
Angel Vargas
re: re: re: re: GE UPS service check
March 13, 2018 09:04
Hello dear, I'm from Venezuela and I need help to take the alarm off the service to a GE SG series 0 80KVA UPS, if you know how to do it I would appreciate it if you could share that information with me, since the maintenance was done to the team but the Alarm does not disappear. Thank you
to rate and post a comment
Attabaa
re: GE UPS service check
November 24, 2016 12:27
You have to configure some of Parameters for said UPS, Set parameter 68 to 20000
to rate and post a comment
Angel Vargas
re: GE UPS service check
March 13, 2018 09:04
Hello dear, I'm from Venezuela and I need help to take the alarm off the service to a GE SG series 0 80KVA UPS, if you know how to do it I would appreciate it if you could share that information with me, since the maintenance was done to the team but the Alarm does not disappear. Thank you
to rate and post a comment
Lenny Place
re: re: GE UPS service check
March 14, 2018 09:44
you need to go to Parameter 68,change the number to 20000
to rate and post a comment
Angel Vargas
re: re: re: GE UPS service check
March 14, 2018 03:13
Thanks for your answer, could you tell me how to change the parameter?
to rate and post a comment
Lenny Place
re: re: re: re: GE UPS service check
March 15, 2018 09:15
please send me an email address and i will send you the installation manual which explains how to access the menus
to rate and post a comment
Angel Vargas
re: re: re: re: re: GE UPS service check
March 18, 2018 11:43
profesor.angelvargas@gmail.com
to rate and post a comment
Tranzrad
re: re: re: re: re: GE UPS service check
January 12, 2019 04:31
Hi Lenny, Do you happen to have any documentation to help me troubleshoot GE SG Series 20kVA UPS. I have all the manuals included on GE Innova 2100/3100/4100IQ SM (under OEM section) but its limited. We have 4 systems which all appear to be error free (all have service light blinking ofcourse) but one of the units does not switch the load to battery (even though the inverter is on and load connected with fresh batteries instaled). UPS just shuts off when removing power. Have checked the wiring and all messages and events, see nothing out of the ordinary. Need documentation else I have to wait 2 weeks for UPS company. We are well versed with UPS systems (and can fully service other OEMS), but GE has extremely limited documentation Thanks in advance for any help you can provide. matt@tranzrad.com
to rate and post a comment
Angel Vargas
re: re: re: GE UPS service check
March 14, 2018 03:13
Thanks for the answer, can you tell me how to change the parameter?
to rate and post a comment
Mkrtich Altunyan
re:re: re: GE UPS service check
October 24, 2019 03:44
Hi Dear. I have a problem with pm. I already replace the capacitors and fans, but wen te ups starting i hav a massage replacw it. Can you help me to tell wich values need to change. And how i can gererate the service key because the service password is don't matching.