GE LightSpeed with Jedi HP60 generator problem

October 04, 2017 09:21

I have a problem with a GE LightSpeed CT scanner.
This system has the Jedi HP60 generator.

the first error is 260118367
Host: On-Board Rotating Processor
An unexpected JEDI State occurred
Jedi Not Ready.

The second error is: 260118260
Host: Jedi
Overpress or blown fuse Error 80-1422

Can someone tell me something sbout this?

Sincerely:

Manuel Brugat/Image Instrumentation, Inc.

Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: GE LightSpeed with Jedi HP60 generator problem

October 05, 2017 09:04

80-1422h- H4 Power IF com. error

Log inor Register

to rate and post a comment

(116)
(10)
(6)

Manuel Brugat

re: re: GE LightSpeed with Jedi HP60 generator problem

October 06, 2017 10:02

Thank you for your answer. I understand there is a problem with the IF board may be. But what is causing it?

Log inor Register

to rate and post a comment

(303)
(47)

CRAIG WAMBURA

re: GE LightSpeed with Jedi HP60 generator problem

October 05, 2017 09:05

Mybe you need to reload the JEDI software

Log inor Register

to rate and post a comment

(116)
(10)
(6)

Manuel Brugat

re: re: GE LightSpeed with Jedi HP60 generator problem

October 06, 2017 10:02

Thank you for your answer. How do i load the JEDI software?

Log inor Register

to rate and post a comment

(303)
(47)

CRAIG WAMBURA

re: GE LightSpeed with Jedi HP60 generator problem

October 06, 2017 02:26

code 80 points to a hardware error.
80-1422 points to overpressure of tube oil or broken fuse.

1. check fuse on IF board
2. check oil pump
3. check communication between CT IF Board and inverter
4. check cabling
5. check for power IF board failure
6. check tube for oil overpressure

Log inor Register

to rate and post a comment

(2)

Douglas Murillo

re: re: GE LightSpeed with Jedi HP60 generator problem

October 11, 2019 10:56

how can i check tube for oil overpressure?

Log inor Register

to rate and post a comment

(331)
(67)

Gabriel Cool

re: re: re: GE LightSpeed with Jedi HP60 generator problem

October 14, 2019 10:17

Check logs

Log inor Register

to rate and post a comment