Over 90 Total Lots Up For Auction at One Location - WA 04/08

CR-85x Error 224AF

(4)

Korvin Kel

CR-85x Error 224AF

June 10, 2019 10:25

Hello. Have a error during the self-test stage when turned on.

Error Code --> 224af
E_CAM_INIT_JAM,
Reason : Input light barrier of cassette module doesn't become free.
Meaning : Cassette jam in cassette module during initialisation.
Cure : Remove cassette.

Meaning : User fault Input of cassette in input buffer.
Cure : Remove cassette and put into buffer with correct orientation.

Of course, there are no cassettes in the receiving buffer, as well as their jams.
Found on this forum the following advice regarding this error:

"Open Cassette module and check the condition of both cassette modules and also check the light barriers.
Check that the light barrier transmitters are secured in their mounts. It might be possible that one of the barriers has fallen out.

Also check the cassette module belt, make sure that the block that is attached to the belt is still there. I've had problems with this block breaking off."

What are the light barrier transmitters and where to find them?
What kind of unit on the cassette belt are we talking about?

Thanks.

Log inor Register

to rate and post a comment

(4)

Korvin Kel

re: CR-85x Error 224AF

June 19, 2019 09:19

We managed to solve this problem. However, there is a new one - the plate is not correctly inserted into the cassette, 2-3 mm to the right than necessary. Is there any other way to get the plate into the cassette a little to the left?

Log inor Register

to rate and post a comment

(4)

Korvin Kel

re: re: CR-85x Error 224AF

June 20, 2019 09:47

With the help of experiments, it was found that this problem is only in the cassette unit 1

Log inor Register

to rate and post a comment

stats
DOTmed text ad

stats
advertisement

Upcoming Auction Events

NorthWest Supply
TIMED
Apr 08, 2024 (99 lots)