Carestream Classic CR problem

May 11, 2020 10:40

Classic CR scanner not extracting the plate from chassis.

Log inor Register

to rate and post a comment

(409)
(82)
(3)

juan pablo amado

re: Carestream Classic CR problem

May 15, 2020 11:42

do you have error code?

Log inor Register

to rate and post a comment

(2)

Kamil Bayram

re: re: Carestream Classic CR problem

May 25, 2020 11:04

951-12021

Log inor Register

to rate and post a comment

(409)
(82)
(3)

juan pablo amado

re: re: re: Carestream Classic CR problem

May 26, 2020 10:47

I can´t find 12021 error,

12012 is:
Description: The PLATE did not move within the correct time.
Cause: A malfunction occurred in one or more of the following components:
• COIL BOARD
• SLOW SCAN AY
Action: 1. Check for an obstruction.
2. Do the checkout for the SLOW SCAN CONTROLLER BOARD A6 and COIL
BOARD A7

Log inor Register

to rate and post a comment

(2)

Kamil Bayram

re: re: re: re: Carestream Classic CR problem

May 27, 2020 11:12

Thank you