M 351 FA_INVERTER_A LARM X-ray generator not available
Switch system off/on Call service if problem is persistent FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.
M 208 XACT_GONE Subsystem error,
system still available It's possible to continue Unexpected X-ray switch off. Check board SHA4 and the generator unit SE. Check cabling.
U have a bad HV tank!! Replace it... Test: fails in High KV!!! Doesnt fail in low KV!!! Replace and calibrate system! U will beed BV scope and philisp Ist!!!
351
-
x ray generator not available Switch system off/on.
FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.
208
Subsystem error, system still available. It's possible to continue
Unexpected X ray switch off. Check board SHA4 and the
generator unit SE. Check cabling.
Hi again Lenny! You seem to be the man to go to about the C Arms. We have a Philips BV Endura error code M050 - System Initialization problem. It happens, we turn off and on several times, and it works for a bit. It is definitely happening more frequently as of recent. Yesterday, it would not take a picture via the hand switch or foot pedal and when it started working again, it only took a picture in Mag1. This morning, it seems to be okay so far.
odilon cardoso
BV Endura Error M351 and M208
February 11, 2016 11:47
BV Endura Error M351 and M208, Can you help me?
to rate and post a comment
Lenny Place
re: BV Endura Error M351 and M208
February 11, 2016 05:53
M 351 FA_INVERTER_A LARM X-ray generator not available
Switch system off/on Call service if problem is persistent FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.
M 208 XACT_GONE Subsystem error,
system still available It's possible to continue Unexpected X-ray switch off. Check board SHA4 and the generator unit SE. Check cabling.
to rate and post a comment
Dean Smith
re: re: BV Endura Error M351 and M208
February 12, 2016 09:01
U have a bad HV tank!! Replace it... Test: fails in High KV!!! Doesnt fail in low KV!!! Replace and calibrate system! U will beed BV scope and philisp Ist!!!
Regards,
Dean
to rate and post a comment
odilon cardoso
re: re: re: BV Endura Error M351 and M208
February 12, 2016 01:51
When put in or more 65kV, ocurre this error, but in 40 45 50 and 60kv no problem.
to rate and post a comment
Lenny Place
re: re: re: re: BV Endura Error M351 and M208
February 12, 2016 03:40
bad mono block
to rate and post a comment
(1)
Kirill Severtsov
re: re: re: re: re: BV Endura Error M351 and M208
June 07, 2018 09:42
How to fix errors M351 and M208
to rate and post a comment
(1)
Kirill Severtsov
re: re: re: re: re: BV Endura Error M351 and M208
June 07, 2018 09:42
How did you fix the errors M208 and M351 of BV Endura 2.3
to rate and post a comment
Basheer Almetwakel
re: re: re: BV Endura Error M351 and M208
June 03, 2021 10:28
How to calibration it please help me to calibration it thanks
to rate and post a comment
DMEDS
re: BV Endura Error M351 and M208
February 12, 2016 09:01
351
-
x ray generator not available Switch system off/on.
FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.
208
Subsystem error, system still available. It's possible to continue
Unexpected X ray switch off. Check board SHA4 and the
generator unit SE. Check cabling.
if you need complete service manuals contact me,
magmesme@gmail.com
Regards
to rate and post a comment
lin
re: re: BV Endura Error M351 and M208
June 12, 2018 08:38
Mr i need the service manuals ,pls help me,send the manuals to the srey123@163.com,thanks
to rate and post a comment
Lenny Place
re: re: re: BV Endura Error M351 and M208
June 12, 2018 11:09
I am sending using myairbridge
to rate and post a comment
lin
re: re: re: re: BV Endura Error M351 and M208
June 13, 2018 07:44
Can you send the manual by srey123@163.com ,thanks
to rate and post a comment
Lenny Place
re: re: re: re: re: BV Endura Error M351 and M208
June 14, 2018 01:44
Sending again
to rate and post a comment
lin
re: re: re: re: re: re: BV Endura Error M351 and M208
June 15, 2018 07:59
Thank you !I have received the manual,the manual in the junk mail.
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: BV Endura Error M351 and M208
June 15, 2018 05:13
Hope it helps you are welcome
to rate and post a comment
Tara
re: re: re: re: re: re: re: re: BV Endura Error M351 and M208
August 19, 2021 12:44
Hi again Lenny! You seem to be the man to go to about the C Arms. We have a Philips BV Endura error code M050 - System Initialization problem. It happens, we turn off and on several times, and it works for a bit. It is definitely happening more frequently as of recent. Yesterday, it would not take a picture via the hand switch or foot pedal and when it started working again, it only took a picture in Mag1. This morning, it seems to be okay so far.
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208
August 20, 2021 10:40
do you see any other error messages? m050 is pretty broad
to rate and post a comment
Tara
re: re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208
August 23, 2021 11:02
No other error messages but the pictures are blurrier than before and we have to use the pulse button to try and get a clearer image on the screen.
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208
August 26, 2021 11:19
Check cabling between of SHA1X3
(stand) and X8 (DFI).
Restart the system?
Problem remains?
DFI or MIDAS defect.
to rate and post a comment
(8)
RobertCow
re: re: re: re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208
September 23, 2021 09:46
try to reload software after check the DFI.
to rate and post a comment