BV Endura Error M351 and M208

odilon cardoso

BV Endura Error M351 and M208

February 11, 2016 11:47

BV Endura Error M351 and M208, Can you help me?

1 person liked this
Log inor Register

to rate and post a comment

Lenny Place

re: BV Endura Error M351 and M208

February 11, 2016 05:53

M 351 FA_INVERTER_A LARM X-ray generator not available
Switch system off/on Call service if problem is persistent FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.


M 208 XACT_GONE Subsystem error,
system still available It's possible to continue Unexpected X-ray switch off. Check board SHA4 and the generator unit SE. Check cabling.

Log inor Register

to rate and post a comment

Dean Smith

re: re: BV Endura Error M351 and M208

February 12, 2016 09:01

U have a bad HV tank!! Replace it... Test: fails in High KV!!! Doesnt fail in low KV!!! Replace and calibrate system! U will beed BV scope and philisp Ist!!!

Regards,

Dean

Log inor Register

to rate and post a comment

odilon cardoso

re: re: re: BV Endura Error M351 and M208

February 12, 2016 01:51

When put in or more 65kV, ocurre this error, but in 40 45 50 and 60kv no problem.

Log inor Register

to rate and post a comment

Lenny Place

re: re: re: re: BV Endura Error M351 and M208

February 12, 2016 03:40

bad mono block

Log inor Register

to rate and post a comment

(2)
(1)

Kirill Severtsov

re: re: re: re: re: BV Endura Error M351 and M208

June 07, 2018 09:42

How to fix errors M351 and M208

Log inor Register

to rate and post a comment

(2)
(1)

Kirill Severtsov

re: re: re: re: re: BV Endura Error M351 and M208

June 07, 2018 09:42

How did you fix the errors M208 and M351 of BV Endura 2.3

1 person liked this
Log inor Register

to rate and post a comment

Basheer Almetwakel

re: re: re: BV Endura Error M351 and M208

June 03, 2021 10:28

How to calibration it please help me to calibration it thanks

Log inor Register

to rate and post a comment

DMEDS

re: BV Endura Error M351 and M208

February 12, 2016 09:01

351
-
x ray generator not available Switch system off/on.
FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.

208

Subsystem error, system still available. It's possible to continue
Unexpected X ray switch off. Check board SHA4 and the
generator unit SE. Check cabling.

if you need complete service manuals contact me,

magmesme@gmail.com

Regards

2 people liked this
Log inor Register

to rate and post a comment

lin

re: re: BV Endura Error M351 and M208

June 12, 2018 08:38

Mr i need the service manuals ,pls help me,send the manuals to the srey123@163.com,thanks

1 person liked this
Log inor Register

to rate and post a comment

Lenny Place

re: re: re: BV Endura Error M351 and M208

June 12, 2018 11:09

I am sending using myairbridge

Log inor Register

to rate and post a comment

lin

re: re: re: re: BV Endura Error M351 and M208

June 13, 2018 07:44

Can you send the manual by srey123@163.com ,thanks

Log inor Register

to rate and post a comment

Lenny Place

re: re: re: re: re: BV Endura Error M351 and M208

June 14, 2018 01:44

Sending again

2 people liked this
Log inor Register

to rate and post a comment

lin

re: re: re: re: re: re: BV Endura Error M351 and M208

June 15, 2018 07:59

Thank you !I have received the manual,the manual in the junk mail.

Log inor Register

to rate and post a comment

Lenny Place

re: re: re: re: re: re: re: BV Endura Error M351 and M208

June 15, 2018 05:13

Hope it helps you are welcome

1 person liked this
Log inor Register

to rate and post a comment

(3)

Tara

re: re: re: re: re: re: re: re: BV Endura Error M351 and M208

August 19, 2021 12:44

Hi again Lenny! You seem to be the man to go to about the C Arms. We have a Philips BV Endura error code M050 - System Initialization problem. It happens, we turn off and on several times, and it works for a bit. It is definitely happening more frequently as of recent. Yesterday, it would not take a picture via the hand switch or foot pedal and when it started working again, it only took a picture in Mag1. This morning, it seems to be okay so far.

Log inor Register

to rate and post a comment

Lenny Place

re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208

August 20, 2021 10:40

do you see any other error messages? m050 is pretty broad

Log inor Register

to rate and post a comment

(3)

Tara

re: re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208

August 23, 2021 11:02

No other error messages but the pictures are blurrier than before and we have to use the pulse button to try and get a clearer image on the screen.

Log inor Register

to rate and post a comment

Lenny Place

re: re: re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208

August 26, 2021 11:19

Check cabling between of SHA1X3
(stand) and X8 (DFI).
Restart the system?
Problem remains?
DFI or MIDAS defect.

Log inor Register

to rate and post a comment

(40)
(8)

RobertCow

re: re: re: re: re: re: re: re: re: re: re: re: BV Endura Error M351 and M208

September 23, 2021 09:46

try to reload software after check the DFI.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Stretcher Rentals

50+ Stretchers in stock for immediate rentals. Contact us for a today for a quote.

DOTmed Certified DOTmed 100 stats
advertisement

Próximos Eventos de Subasta

NorthWest Supply
TIMED
Apr 08, 2024 (99 lots)