Service manual for Abbott Cell-Dyn 1700 CBC system
June 30, 2011 05:22
I am looking for a service manual for a Cell-Dyn 1700. I am working on it for a buddy who runs a free clinic, trying to track down "Pressure Level Time Out" and "Vacuum Level Time Out" errors.
Re: Service manual for Abbott Cell-Dyn 1700 CBC system
July 09, 2011 05:06
Pressure level timeout means that the pressure pump was unable to satisfy the pressure sensor (0.5PSI) within 45 seconds. This means there is a leak or the pump is not running at all.
Vacuum level timeout means that the vacuum pump was unable to satisfy the vacuum sensor (8" Hg) within 45 seconds. This also means there is a leak or the pump is not running at all.
If you get both indications, I would bet that the pumps are not running at all. Remember that a single tubing disconnected or removed from a pinch valve can cause this, as well as rusty solenoids that don't close (usually near the bottom of the front panel).
email me if you need more help. The service manual would be of no help diagnosing this.
(10)
Ty Murphy
Service manual for Abbott Cell-Dyn 1700 CBC system
June 30, 2011 05:22
I am looking for a service manual for a Cell-Dyn 1700. I am working on it for a buddy who runs a free clinic, trying to track down "Pressure Level Time Out" and "Vacuum Level Time Out" errors.
Thx
to rate and post a comment
(10)
(15)
Stephen Saftler
Re: Service manual for Abbott Cell-Dyn 1700 CBC system
July 09, 2011 05:06
Pressure level timeout means that the pressure pump was unable to satisfy the pressure sensor (0.5PSI) within 45 seconds. This means there is a leak or the pump is not running at all.
Vacuum level timeout means that the vacuum pump was unable to satisfy the vacuum sensor (8" Hg) within 45 seconds. This also means there is a leak or the pump is not running at all.
If you get both indications, I would bet that the pumps are not running at all. Remember that a single tubing disconnected or removed from a pinch valve can cause this, as well as rusty solenoids that don't close (usually near the bottom of the front panel).
email me if you need more help. The service manual would be of no help diagnosing this.
Steve
to rate and post a comment
(10)
Ty Murphy
Re: Service manual for Abbott Cell-Dyn 1700 CBC system
July 11, 2011 01:12
Many thanks, Steve!
to rate and post a comment
RASSHESH P DANI
Re: Service manual for Abbott Cell-Dyn 1700 CBC system
August 16, 2011 08:36
do u have service manual of abott cd-1700? i am a freelancer, in bombay, india.
- rassheshpdani@gmail.com
to rate and post a comment
RASSHESH P DANI
Re: Service manual for Abbott Cell-Dyn 1700 CBC system
September 01, 2011 11:40
dear stephan,
one cd-1700 shows intializing error - sequence # 1071. what are the possibilities?
- dani r [bombay, india]
to rate and post a comment